Optical density, normalization, and growth rates
Background on optical density​
Light passing through a culture behaves differently than light passing through a clear solution. Instead of being absorbed, light is scattered by the cells in suspension. This scattered light is measured as optical density (OD). As turbidity increases, more scattering occurs, resulting in a higher OD reading.
Looking to calibrate for OD600 readings? Visit this page.
Normalization​
Due to manufacturing variation (LED strength and sensor sensitivity), raw OD readings can't be compared directly across devices. Instead, we use the normalization technique described below:
A series of initial OD readings are averaged to produce a reference value (baseline). New OD readings after the reference value are normalized using the following simple equation:
latest  ODreference  OD=normalized  OD\frac{\normalsize \text{latest\thickspace OD}}{\normalsize \text{reference\thickspace OD}}=\text{normalized\thickspace OD}
For example:
Pioreactor nameReference ODLatest OD
Pioreactor10.0300.033
Pioreactor20.0100.015
It's difficult to compare raw OD readings since the starting values are different. However, if we normalize using the above equation:
Pioreactor1​
0.0330.030=1.1\frac{\small 0.033}{\small 0.030}=1.1
Culture has grown by 1.1x.
Pioreactor2​
0.0150.010=1.5\frac{\small 0.015}{\small 0.010}=1.5
Culture has grown by 1.5x.
We can more accurately compare culture growth using these ratios as opposed to using the raw OD values.
Because of the way we defined normalized optical density, it has the following easy interpretation: it's the multiplicative amount the culture has changed by. So if the normalized OD is 2.0, the culture has doubled its initial concentration, i.e. doubled the population since the volume is fixed. This interpretation also maps to traditional OD600 measurements: if your initial sample has OD600 equal to 0.45, then a normalized OD of 2.0 is approximately an OD600 of twice that, or 0.90.
Blanking​
While basic normalization accounts for initial OD differences, it does not consider the optical density of the media itself. For a more accurate growth rate calculation, you can blank your sample.
Blank uncalibrated vials when the uninoculated media already has visible or measurable turbidity. By blanking, you subtract that media background and observe the OD change associated with the microorganism of interest.
Per-experiment blanking is for uncalibrated OD measurements. Do not record a blank while an OD calibration or fused OD estimator is active; the Pioreactor UI disables this incompatible combination.
If the measured signal later falls below the stored blank, Pioreactor does not publish a negative OD reading. If this continues, clear the experiment's blank or record it again using the current media and optical setup.
As an example, let's consider the same data as above, but this time we have information on the blank ODs:
Pioreactor nameBlank OD (media)Reference OD (media with culture)Difference (reference - blank)Latest OD
Pioreactor10.0250.0300.0050.033
Pioreactor20.0050.0100.0050.015
We can now subtract the blank values from both the latest OD and reference OD values:
latest  OD−blank  ODreference  OD−blank  OD=blanked  normalized  OD\frac{\normalsize \text{latest\thickspace OD}-\text{blank\thickspace OD}}{\normalsize \text{reference\thickspace OD}-\text{blank\thickspace OD}}=\text{blanked\thickspace normalized\thickspace OD}
Pioreactor1​
0.033−0.0250.030−0.025=1.6x\frac{\small 0.033-0.025}{\small 0.030-0.025}=1.6x
Culture has grown by 1.6x.
Pioreactor2​
0.015−0.0050.010−0.005=2x\frac{\small 0.015-0.005}{\small 0.010-0.005}=2x
Culture has grown by 2x.
By accounting for the OD of the blank media, we are able to calculate a more accurate growth rate.
Growth rate​
We inoculated two vials with a drop of re-hydrated yeast, and tracked their growth at temperatures 27°C and 35°C. The UI shows the following normalized optical density chart:
From normalized optical density, the UI also computes an implied growth rate. The relationship between the implied growth rate, gr\text{gr}, and the normalized optical density, nOD\text{nOD}, is exponential and can be written as:
nOD(t)=exp⁡(∫0tgr(s)ds)\text{nOD}(t) = \exp{ \left( \int_0^t \text{gr}(s)ds \right)}
We can rewrite this equation in terms of the growth rate:
gr(t)=nOD′(t)nOD(t)\text{gr}(t)= \frac{ \text{nOD}'(t)} {\text{nOD}(t)}
This states that the growth rate is the rate of change of the culture size, normalized by the culture size.
This rate helps compare the state of your culture under different conditions (temperature, media, etc.).
These graphs can be interpreted in four phases:
  • The lag phase: Little observed growth, but high cell activity. Cells acclimate to a nutrient-rich environment and prepare for division.
  • Exponential (or log) phase: Cells divide and double at a roughly constant rate. Generation times depend on species and conditions.
  • Stationary phase: Eventually the growth of cells reaches a plateau as nutrients are used up and waste products accumulate. At this point, the number of dividing cells will equal the number of dying cells.
  • Decline (or death) phase: As nutrients are depleted, cell growth slows while cell death increases.
Here's how these phases would apply to our graphs, focusing on the 35°C vial:
Some things to note:
  • The lag phase can be detected easily in the growth rate graph, as the rate is stable and doesn't begin increasing until a bit after 6 PM. This is not easily determined in the nOD graph, since at this point the overall turbidity of the culture is low.
  • The exponential phase occurs when the growth rate is high/increasing.
  • When the culture reaches the stationary phase, growth rate drops to 0 since the culture is no longer growing in size. The turbidity is constant.
  • The decline phase is not represented in the graphs above, as yeast remain in the stationary phase over many days.
(Optional) Blanking​
When working with small amounts of a microorganism, or when using very turbid media, you can obtain more accurate growth rates by blanking the vial before inoculation.
How to record a blank #
  1. Insert your sterile vial containing media into the Pioreactor before inoculating with your species of interest.
  2. In the UI, click the Pioreactors tab on the left-hand menu, and choose one of the active pioreactors.
  3. Select Calibrate, and under the Blanks tab, click Start. The Pioreactor will now record the optical density of the blank vial.
  4. Repeat for all the Pioreactors to be used.
    • A notification will appear when a Pioreactor has finished blanking.
  5. You can now inoculate your vials and begin your experiment.